Chiral AGP and Chiral HSA column evaluation
I need to evalute chiral AGP and chiral HSA column, but i don't know procedure testing. So i need some procedures testing and acceptance value for reference. Can you help me?
If you will be using a column for a specific method, the best way is to run a "mini system suitability" for that method. You don't have to run replicates, but just a single check sample to verify that retention, plates, tailing, etc. are OK.
If you are getting the columns for "general purpose" use, the best way is to re-run the manufacturer's test.