Surrogate recovery of 4 BFB
Posted: Wed Jun 19, 2013 4:38 pm
by aes1311
Using Archon autosampler, p&t OI 4560,GC 5980 series II and MSD 5972 having issues with erratic BFB recoveries. Clean matrix recovery ok , however introduce high concentration sample or 200 ppb level cal std and BFB recovers high . MSD source was clean recently.
Re: Surrogate recovery of 4 BFB
Posted: Thu Jun 20, 2013 3:42 pm
by James_Ball
Larger amounts of Methanol from standards can cause sensitivity changes such as this though I have normally saw it happen with lighter early eluting analytes more. I used to have the same problem with 1,2-Dichloroethane-d4 surrogate.
Are you using a dry purge in the cycle?
What is desorb time and flow rate?
Are you using the regular K&M electron multiplier?
When was the last time you changed the fluid in the diffusion pump?
Do you have a molecular sieve moisture trap at the rough pump?
A few things I am thinking of that all have an effect on volatiles as I have seen in the past.
Posted: Fri Jun 21, 2013 8:52 pm
by aes1311
no dry purge (using #10 trap)
desorb time 1.0 min @10 mL/minute
yes K&M electron multiper
pump fluid replace when the source was clean
yes to moisture trap
Posted: Mon Jun 24, 2013 6:51 pm
by James_Ball
The Tenax, Silica Gel, Charcoal #10 trap tends to hold a lot of water which gets desorbed onto the MS, the water loading of the source on the 5972 can cause some problems like this.
We moved away from this trap many years ago and went to the VoCarb3000 which can use the dry purge to remove some of the water, it does help some.
Is this going through a split injection port? If so you may want to increase the desorb flow/split ratio to also introduce less water and methanol into the system. The higher flow rate will also help pull things off the trap better.
The one thing we did to improve stability for instruments like this was switch to the ETP ribbon style Electron Multiplier, they seem to be less prone to having sensitivity changes as you load the instrument up with organics. One thing that I used to notice the the K&M was when you get hit with something such as gasoline at high levels in a sample was that the sensitivity would increase for several injections afterwards before returning to normal, similar problems with high level standards.
Posted: Mon Jul 01, 2013 8:18 pm
by aes1311
thanks for assistance, the problem was a bad lot of refurbished filaments. Replace with agilent filaments and problem corrected.