Surrogate recovery of 4 BFB
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
5 posts
Page 1 of 1
Using Archon autosampler, p&t OI 4560,GC 5980 series II and MSD 5972 having issues with erratic BFB recoveries. Clean matrix recovery ok , however introduce high concentration sample or 200 ppb level cal std and BFB recovers high . MSD source was clean recently.
-
- Posts: 3477
- Joined: Mon Jan 07, 2013 8:54 pm
Larger amounts of Methanol from standards can cause sensitivity changes such as this though I have normally saw it happen with lighter early eluting analytes more. I used to have the same problem with 1,2-Dichloroethane-d4 surrogate.
Are you using a dry purge in the cycle?
What is desorb time and flow rate?
Are you using the regular K&M electron multiplier?
When was the last time you changed the fluid in the diffusion pump?
Do you have a molecular sieve moisture trap at the rough pump?
A few things I am thinking of that all have an effect on volatiles as I have seen in the past.
Are you using a dry purge in the cycle?
What is desorb time and flow rate?
Are you using the regular K&M electron multiplier?
When was the last time you changed the fluid in the diffusion pump?
Do you have a molecular sieve moisture trap at the rough pump?
A few things I am thinking of that all have an effect on volatiles as I have seen in the past.
The past is there to guide us into the future, not to dwell in.
-
- Posts: 8
- Joined: Tue Jun 18, 2013 7:41 pm
no dry purge (using #10 trap)
desorb time 1.0 min @10 mL/minute
yes K&M electron multiper
pump fluid replace when the source was clean
yes to moisture trap
desorb time 1.0 min @10 mL/minute
yes K&M electron multiper
pump fluid replace when the source was clean
yes to moisture trap
-
- Posts: 3477
- Joined: Mon Jan 07, 2013 8:54 pm
The Tenax, Silica Gel, Charcoal #10 trap tends to hold a lot of water which gets desorbed onto the MS, the water loading of the source on the 5972 can cause some problems like this.
We moved away from this trap many years ago and went to the VoCarb3000 which can use the dry purge to remove some of the water, it does help some.
Is this going through a split injection port? If so you may want to increase the desorb flow/split ratio to also introduce less water and methanol into the system. The higher flow rate will also help pull things off the trap better.
The one thing we did to improve stability for instruments like this was switch to the ETP ribbon style Electron Multiplier, they seem to be less prone to having sensitivity changes as you load the instrument up with organics. One thing that I used to notice the the K&M was when you get hit with something such as gasoline at high levels in a sample was that the sensitivity would increase for several injections afterwards before returning to normal, similar problems with high level standards.
We moved away from this trap many years ago and went to the VoCarb3000 which can use the dry purge to remove some of the water, it does help some.
Is this going through a split injection port? If so you may want to increase the desorb flow/split ratio to also introduce less water and methanol into the system. The higher flow rate will also help pull things off the trap better.
The one thing we did to improve stability for instruments like this was switch to the ETP ribbon style Electron Multiplier, they seem to be less prone to having sensitivity changes as you load the instrument up with organics. One thing that I used to notice the the K&M was when you get hit with something such as gasoline at high levels in a sample was that the sensitivity would increase for several injections afterwards before returning to normal, similar problems with high level standards.
The past is there to guide us into the future, not to dwell in.
-
- Posts: 8
- Joined: Tue Jun 18, 2013 7:41 pm
thanks for assistance, the problem was a bad lot of refurbished filaments. Replace with agilent filaments and problem corrected.
5 posts
Page 1 of 1
Who is online
In total there are 36 users online :: 0 registered, 0 hidden and 36 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 36 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 36 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science