High base line in EI GC/MS
Discussions about GC and other "gas phase" separation techniques.
4 posts
Page 1 of 1
I have a trouble with high baseline results. Do you know why this is happened?
-
- Posts: 960
- Joined: Thu Apr 18, 2013 2:10 pm
Without more details, you're not going to get very good answers. Does your MS give you a reasonable tune?
-
- Posts: 6
- Joined: Thu May 30, 2013 5:02 pm
yes I got a normal tune result.
I'm thinking about something in the GC/MS methods that make it high.
Do you have any advice about this?
I'm thinking about something in the GC/MS methods that make it high.
Do you have any advice about this?
-
- Posts: 1680
- Joined: Sat Aug 23, 2008 12:04 am
This can be air, column bleed, a data acquisition parameter like a detection threshold. If you have a sample introduction system other than liquid injection with a syringe, it could be a malfunction with that injection system. If yoy have a split to some other components in the GC oven, it could be related to them. And, even though the good tune would suggest against it, it could be an electronic problem.
Given that each kind of instrument has different properties (and problems common to that instrument), what kind of instrument (make and model) and what version software are you using?
Waht is the instrument configuration?
What was the background level when all was well and what has happened to the instrument since. What is the background level now? What was the scan range then and what is it now?
How do air/water checks look?
Do you have a full scan of the background? Pictures really help - there are instructions for posting.
Given that each kind of instrument has different properties (and problems common to that instrument), what kind of instrument (make and model) and what version software are you using?
Waht is the instrument configuration?
What was the background level when all was well and what has happened to the instrument since. What is the background level now? What was the scan range then and what is it now?
How do air/water checks look?
Do you have a full scan of the background? Pictures really help - there are instructions for posting.
4 posts
Page 1 of 1
Who is online
In total there are 16 users online :: 2 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Baidu [Spider], Majestic-12 [Bot] and 14 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Baidu [Spider], Majestic-12 [Bot] and 14 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science