Although I've used this site as a reference many times, this is my first post. Hopefully, an expert/veteran has some advice for me.
So, I've been battling this issue for about 3 weeks now and cannot figure it out. We measure a drug in 80/20 MethOH/H2O. After letting the our Agilent 1100 sit for about a month, I am getting a front/shoulder on the peak. I cannot figure out what the problem is. I have rebuilt pumps (seals and new inlet and outlet valves), new frit/cap, new column, new lines. I have tried measuring other compounds with other mobile phase and fronting occurs with different samples and different mobile phases. I've created new methods and it persists. I've bypassed degasser and damper/pressure sensor and it persists. I've rebuilt rheodyne valve also. I've run the mobile phase on individual channels and premixed on each pump and it persists. I have replaced injector with a spare and it persists. I've replaced inlet filters also.
Any advice?
I'm about to reinstall Chemstation because Agilent tech told me that sometimes methods will continiously get "corrupt" and induce micro pressure flunctuations that don't show up on pressure readout.
System details: Solvent => vacuum degasser (G1322A)=>binary pump (G1312A)=> 3rd party injector/oven (8125 rheodyne valve)=> Agilent C8 guard column=> Agilent C18 column=> UV/Vis DAD (G1314A)=> 100psi backpressure regulator