8270 SVOA Internal Standard from Neat Compounds

Discussions about sample preparation: extraction, cleanup, derivatization, etc.

3 posts Page 1 of 1
Has anyone ever made their own Internal STD for EPA8270 from neat compounds? I'm trying to make a 4000ppm SV Internal standard solution in DCM and everything goes into solution fine except for Perylene-d12. Even after 25 minutes of sonication, I still have some undissolved powder in the bottom of my volumetric flask. A last ditch effort seems to be heating the solution up to almost boiling and hoping the remaing solid dissolves, but I'm worried about losing some of my low boilers (1,4-DCB-d4, Naph-d8). Any suggestions? Thanks for your time.
Our folks do it routinely - they tell me they just use an ultra-sonic bath. That being said, I often see lousy variations in the internal standard mixes they make up, so I bet they have more trouble with this issue than they will admit.
Thanks Yama. That's exactly what I've been doing but it takes about 20-30 minutes of sonication and a good half day sitting at room temperature for all the Perylene-d12 to dissolve back into solution. Guess I should just be happy it does. :)
3 posts Page 1 of 1

Who is online

In total there are 6 users online :: 0 registered, 0 hidden and 6 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 6 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry