last time I posted i was directed to another podt however my investigations into this have resulted in an extremely sore head.
im working on an aquity H-class using an FLR detector
under gradient elution (5-95% over 5mins + reequilibration) on a 1.7um kinetex column at 285/350nm the compound is burprenorphine.
mobile phase: Acn:H2O (0.1% TFA) and working to a nomial calibration line of 0.2ppb-20ppb.
ultimate goal: extraction from plasma
my peak is dwarfed by this interference which starts at approx 2.5mins and increases with the increasing solvent strength.
i have tried everything in my knowledge to remove the interference including a system.wash with 30% phosphoric acid + pH neutralization. this lowered the interference ftom around 500EU to about 50 however my top calibration spike response (in solution is about 20). i have tried running the gradient (with no injection) with bottled lc water, milliq water, different brands of ACN and with/without Tfa but the interference is still there! even flushing in 100% Acn doesnt remove it!
unfortunatly management says chemistry isnt their job so get it sorted. with no managment direction and only 2.5yrs of chromatography experience I feel i am at a loss.
the only thing I can think of is a mixing issue but its just ACN and water!
I have finally achieved the sensitvity required (in solution/unextracted) isocratically but the sensitivity is half of the gradient and extracted samples are resulting in interfereneces of their own, therefore i would prefer gradient.
I have also (with absolutly no success) tried 2 dervitisation techniques
dansyl chloride and hexacyanoferrate) using adapted papers. these used diffefent wavelengths and the interference dissapeared.
With my neck on the line and a resounding NO for the purchase of an MS/MS system i am truely begging for help. not only for my job but for my sleep and sanity!
Hope someone can help an amateur chromatographer at the end of his tether
thanks in advance
CM