How critical is the pH of the sample in HPLC analysis. I am doing lactic acid analysis. Sample prep includes a hydrolysis step with NaOH (high pH), neutralization with H2SO4 conc to lower the pH to get the acid in COOH form. If the pH of the sample gets very low (about 1 or less, is this issue for the column’s life? Mobile phase is water acidified with H2SO4 to pH 2, injection volume 20-50 microliters, flow 1.0 mls/min. Column is Apollo C18 silica from Alltech Grace (supplier says is fine in pH range 1-8). I think I am fine, but would like a “second opinion
. As long as we do not charge health care fees!
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
Assuming you're using a "conventional" column (something in the 4 or 4.6 mm id class) that 20-50 microliters gets diluted with mobile phase fairly quickly. If you want to be doubly certain, put in a guard cartridge and let *that* take the hit.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 26
- Joined: Thu Oct 13, 2011 4:41 pm
Thanks Tom, you are the BEST!!!Assuming you're using a "conventional" column (something in the 4 or 4.6 mm id class) that 20-50 microliters gets diluted with mobile phase fairly quickly. If you want to be doubly certain, put in a guard cartridge and let *that* take the hit.
-
- Posts: 4
- Joined: Mon Oct 13, 2014 11:00 am
Sounds good.. such a nice and helping nature you have. Keep helping .. god bless..
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
Mag - we validated a similar test procedure for GLP.
We added 10 ml 0.5 N NaOH to saponify all lactic acid esters, then added 10 ml 0.5 N H2SO4 to neutralize remaining NaOH and to drop pH of the sample. We also used dilute H2SO4 as the mobile phase.
By the way, we also used Na lactate as the standard, as lactic acid solutions (such as USP grade) all have a certain amount of internal ester; this is why the USP procedure for lactic acid content uses saponification and back titration.
We added 10 ml 0.5 N NaOH to saponify all lactic acid esters, then added 10 ml 0.5 N H2SO4 to neutralize remaining NaOH and to drop pH of the sample. We also used dilute H2SO4 as the mobile phase.
By the way, we also used Na lactate as the standard, as lactic acid solutions (such as USP grade) all have a certain amount of internal ester; this is why the USP procedure for lactic acid content uses saponification and back titration.
5 posts
Page 1 of 1
Who is online
In total there are 21 users online :: 2 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], Google [Bot] and 19 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], Google [Bot] and 19 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science