I'm having an issue with peak tailing on an IC, that I can't seem to resolve. It is a Dionex ICS-2000 with a KOH eluent generator cartridge, being used to analyze anions. The tailing is most noticeable on fluoride, nitrite, and ortho phosphate. So far I've replaced the columns, both guard and analytical, didn't fix it (I took them apart to look at the packing and replaced the frits, no dead space was observed.) I've tried a different suppressor from another system, which was not experiencing the same issue, and this also did not correct the problem. I've tried bypassing the carbonate removal device, also doesn't appear to be the problem. There has also been complete preventative maintenance performed on the AS autosampler by Dionex, replacing the needle assembly, transfer line, etc.
I would really appreciate any suggestions. I feel like there has to be some dead space somewhere in the system, just not sure where it's coming from. The system is still usable for the most part, but the problem is really exaggerated at low concentrations for said anions.
By the way, I'm using AS-18 and AG-18 2mm microbore columns and isocratic elution.