recalibration required after inlet maintenance?

Posted: Fri Apr 19, 2013 2:09 pm
by jenniferd
Hi,

For methods 8260/624 and 8270/625, is it required to recalibrate after inlet maintenance (new liner, septum, gold seal, etc.)?

Thanks!

Re: recalibration required after inlet maintenance?

Posted: Fri Apr 19, 2013 6:41 pm
by Yama001
It has not been required in any place I have worked - it is not considered to be a major change. A continuing cal check is all that is normally required.

Re: recalibration required after inlet maintenance?

Posted: Fri Apr 19, 2013 7:28 pm
by James_Ball
As long at your continuing calibration check passes it should not be necessary to recalibrate.