For methods 8260/624 and 8270/625, is it required to recalibrate after inlet maintenance (new liner, septum, gold seal, etc.)?

Thanks!
It has not been required in any place I have worked - it is not considered to be a major change. A continuing cal check is all that is normally required.
As long at your continuing calibration check passes it should not be necessary to recalibrate.