Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
what is shoulder on the front side of the peak.I have a Luna C18(2) column that has worked beautifully through about 3 months of heavy use ~ 30 hrs/wk, but recently, the backpressure is ~20 bar higher than normal and many of the peaks have a shoulder on the front side. At first, I assumed that I was just injecting too much sample, but the shoulder was still there at 1/2 and 1/4 the sample concentration. I thought it might have been a void space or clog in the guard column that caused this, but I tried running some of the same samples without a guard column and got the same results. Also I tried reverse-flow cleaning w/ 1:1 ACN/THF, and that had no effect. Is there anything else I can do that might fix my column? Thanks in advance,
Chris