I have an important question to the community regarding HPLC-MS-MS quantitative analysis using a C18 column. (2.1x100x3micron)
Upon running a sample at 0.5mL/min in 10:90 AQ:ORG with the column in place I Observed a retention time of ~0.5min
While maintaining the flow rate at 0.5mL/min in a 50:50 AQ:ORG with the column in place I observed a new retention time of ~1.7min
The compound is definitely retained, but slightly with lower affinity due to higher organic content. The question is , does a problem arrise if too short a retention time is used, or is this method using the 1.7min RT acceptable?
I am a bit worried due to the small retention, but the calibration curves are consistent and the quantitation is consistent batch to batch.
The elution is isocratic, and an appropriate equilibration and run time is used in every sample.
What is an acceptable short retention time?
Thanks for the help!