I am getting a "lost prime" message with our Waters Alliance 2690 HPLC system. There is no mobile phase coming out in dry prime. I am new to trouble shooting. Has anybody else experienced this problem and if so would you please tell me how to solve it? Thanks!

Hi

We've had similar problems with our 2690's, especially with ACN as mobile phase. An easy way of getting the flow back is by cleaning your check valves. We normally clean our check valves by sonnicating it in MEOH for about 10 min.

Hope this helps.
Mike

It did work!!!

Thank you very much!
Hi

This is normally a quick solution and works in most cases. If it does not solve the problem then you might need to change your check valves. Waters makes a "performance check valve" which I believe is better than the normal check valve. Try this, one of the suggestions will solve your probelm.

Cheers
Mike
4 posts Page 1 of 1

Who is online

In total there are 8 users online :: 1 registered, 0 hidden and 7 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 7 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry