						I know its not the best thing for column longevity but how much damage an I likely to cause to a HP35 column if I inject samples in a water matrix directly onto it? Is there any thing I can do to prevent problems? Would a pre-column do any good?
					
									- 
								
 - Posts: 927
 - Joined: Mon Feb 13, 2006 6:28 pm
 
It largely depends on starting temperature of the oven.I know its not the best thing for column longevity but how much damage an I likely to cause to a HP35 column if I inject samples in a water matrix directly onto it?
If water won't condense in the column it practically won't harm.
- 
								
 - Posts: 880
 - Joined: Fri Jan 04, 2008 3:03 pm
 
						If shooting water, cool down the injection port to just over 100. Maybe 125 or so, otherwise you run the risk of greatly exceeding the volume of your liner.
					
									- 
								
 - Posts: 927
 - Joined: Mon Feb 13, 2006 6:28 pm
 
Yep, and keep injection volume as low as possible. However it has nothing to do with column lifetime.If shooting water, cool down the injection port to just over 100. Maybe 125 or so, otherwise you run the risk of greatly exceeding the volume of your liner.
- 
								
 - Posts: 102
 - Joined: Mon Jan 10, 2011 1:38 pm
 
						Most siloxanes can handle water, but be careful.  The higher the polarity, the more sensitive phases get for hydrolization. There are backbone stabilized phases out there that have even bettre tolerance (35 Sil MS).
For any phase, water is a problem if it gets in the column when it is HOT. (like via the carrier gas or a leaking fitting/septum).
Water as matrix is usually not a problem as the water peak elutes at low oven temperature. Be aware that injection and detection part however, can develop activity, depending on temperature used.
Water evaporation is a bigger issue because of expansion volume. Use 4-5mm ID liners. Be careful with wool. We generally recommend the cyclosplitter or the double gooseneck for inejcting water samples
jaap de zeeuw, Restek corporation
									For any phase, water is a problem if it gets in the column when it is HOT. (like via the carrier gas or a leaking fitting/septum).
Water as matrix is usually not a problem as the water peak elutes at low oven temperature. Be aware that injection and detection part however, can develop activity, depending on temperature used.
Water evaporation is a bigger issue because of expansion volume. Use 4-5mm ID liners. Be careful with wool. We generally recommend the cyclosplitter or the double gooseneck for inejcting water samples
jaap de zeeuw, Restek corporation
			5 posts
							Page 1 of 1
					
	Who is online
				In total there are 9 users online :: 0 registered, 0 hidden and 9 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
 
		
Users browsing this forum: No registered users and 9 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 9 guests
Latest Blog Posts from Separation Science
    Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
    
    
    Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
    
    
- Follow us on Twitter: @Sep_Science
 - Follow us on Linkedin: Separation Science
 