What happens if you inject *two* MeCl2 blanks after a sample? Is the amount of carryover in the second blank 3% of 3% (i.e. 0.09%) or is it higher?
Next question: do you see the same carryover when your run standards (I'm assuming you're trying to quantitate)?
If you see the same % carryover with standards, and if the second blank is very low (under 0.1%), then the easiest fix might be to simply run a MeCl2 blank in between each run as a "cleanup". Your recovery will only be about 97%, but if the standards behave the same way, that will cancel out.
If the carryover in the second blank is much higher (say 1% or so), that suggests that you have stuff adsorbing *somewhere* in the system that will come off in MeCl2 but not in DMSO. If you have a ternary or quaternary gradient system, you might add a final gradient segment where you switch to a MeCl2/iPrOH blend. Again, less than 100% recovery is a problem, but if the standards behave the same way it might be OK.
If you don't see the carryover with standards, then quantitation will be difficult and you will have to explore a different sample prep.
A big part of your problem may be that you are injecting MeCl2 into a highly aqueous system. One work-around might be to add some iPrOH as a co-solvent before you inject (and double the injection volume if detectability is an issue).
Next question: do you see the same carryover when your run standards (I'm assuming you're trying to quantitate)?
If you see the same % carryover with standards, and if the second blank is very low (under 0.1%), then the easiest fix might be to simply run a MeCl2 blank in between each run as a "cleanup". Your recovery will only be about 97%, but if the standards behave the same way, that will cancel out.
If the carryover in the second blank is much higher (say 1% or so), that suggests that you have stuff adsorbing *somewhere* in the system that will come off in MeCl2 but not in DMSO. If you have a ternary or quaternary gradient system, you might add a final gradient segment where you switch to a MeCl2/iPrOH blend. Again, less than 100% recovery is a problem, but if the standards behave the same way it might be OK.
If you don't see the carryover with standards, then quantitation will be difficult and you will have to explore a different sample prep.
A big part of your problem may be that you are injecting MeCl2 into a highly aqueous system. One work-around might be to add some iPrOH as a co-solvent before you inject (and double the injection volume if detectability is an issue).
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 15
- Joined: Fri Mar 29, 2013 7:43 pm
...
3 posts
Page 1 of 1
Who is online
In total there are 30 users online :: 0 registered, 0 hidden and 30 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 30 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 30 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science