Easy question concerning a chiral separation

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
I feel a little dumb asking this since I think I should know the answer, but since my anonymity is guaranteed I'm going to ask anyways.

We have two sets of positional isomers that I have to separate. The two sets are identical in structure except that one set of R and S isomers has a NO2 group and the other set has been reduced to NH2.

The NO2 isomers were separated using a Chiracel-OJ column with conditions 80:20 hexane:IPA, sample diluent EtOH, column temp. 40°C, flow rate 1.0 ml/min, back pressure 40 bar. The isomers are baseline resolved with very good looking sharp gaussian peaks.

Using the exact conditions for the NH2 isomers, the peaks were resolved but extremely broad in shape. They're not taling to be clear, just very low plates. What is the mechanism for this? I'm aware that compounds that are strongly retained will tail, but tailing is 1.0 here. The peaks are just very very broad. 6-7 minutes wide.

If I was trying to deal with a tailing problem, I could add DEA, but I don't think I'm ever had to address poor efficiency before. thanks for any suggestions.
What are the retention times, respectively? Maybe just add more alcohol so they come out sooner and more narrow.
It didn't occur to me when I posted that it could be carryover since I never have this problem until now. Will try 60:40 Hexane:IPA and see what happens. thx
3 posts Page 1 of 1

Who is online

In total there are 28 users online :: 4 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Amazon [Bot], Google [Bot], Semrush [Bot] and 24 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry