Storage of HPLC column

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
Hi,
A HPLC column was flushed with 5%Acetonitrile in H2O after analysis using a buffer salt mobile phase.
What impact would there be on the column if it was left in 5%Acetonitrile in H2O for 4 days?
(Note: The manufacturer's recommended storage solvent for this column is methanol.)
Thanks!
The column might die 2 days sooner than it would if it had been stored in methanol.

It's kind of like the old days of photographic film (remember that?). It was recommended that long term storage be done in the refrigerator, but the film did not self-destruct while sitting on a shelf at room temperature.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
While I seldom find myself wanting more after reading one of TP's posts (as he has forgotten more LC lore than I shall ever learn), I would qualify the above with an "It depends on the column" as there are a few columns that will die if subjected to the "wrong" storage solvents (are you listening, X-Bridge users and manufacturer)?
Thanks,
DR
Image
3 posts Page 1 of 1

Who is online

In total there are 17 users online :: 1 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Majestic-12 [Bot] and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry