I sometimes have to clean the ESI source of 6410 and I just leave the instrument in standby. In case, I have to clean the capillary, do I have to turn off the system or just leave it in standby ?
Thanks.
Advertisement
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
Duno whats better "gunk" or this :To clean the cappillary I shut down the MS
You could do the second suggestion of pepter but I would not recommend it
as you just transfer all the gunk into the ms
I usually clean with isoprop
Using a thin wire with a small loop and some cottonwool drenged in isoprop trapped in the loop
But be careful not to use to much cotton wool
You don't want to get some stuck in the cappillary
1) tuning only takes 10 to 15 minutes1) switch off ms means you need to tune it...
2) switch off ms means you waste time becuase you need to wait for detector stabilization after switching it on
3) switch off means you stop the turbo pump...every stop and run shorther life time of this expensive part
4) when you pulling cappilary out and put it back you scratch the platinum ends...
5) wire loop cottonwool ... worst idea everyou think that they write this way of cleaning because they want you to save time, cash, keep parts longer ... nah.
You tune machine to be sure that your detector is calibrated properly and mass indication is good. When you switching of detector you "destabilizate" it additionally during some maintance you can affect ionization area/ions pathway etc.Thank you all for the help.
Now that I can see I have another question. You said that, when ever shutting off the instrument, the tune must be re-run. Normally we do not tune after shutting off. Is that really a problem ?
You mean even when the power is off, and on again (because of the lines) and I still have to tune the instrument ?You tune machine to be sure that your detector is calibrated properly and mass indication is good. When you switching of detector you "destabilizate" it additionally during some maintance you can affect ionization area/ions pathway etc.Thank you all for the help.
Now that I can see I have another question. You said that, when ever shutting off the instrument, the tune must be re-run. Normally we do not tune after shutting off. Is that really a problem ?
Depends what you do. Here is what we do in our labYou mean even when the power is off, and on again (because of the lines) and I still have to tune the instrument ?You tune machine to be sure that your detector is calibrated properly and mass indication is good. When you switching of detector you "destabilizate" it additionally during some maintance you can affect ionization area/ions pathway etc.Thank you all for the help.
Now that I can see I have another question. You said that, when ever shutting off the instrument, the tune must be re-run. Normally we do not tune after shutting off. Is that really a problem ?
Thanks for your help.Cleaning the ion source and accessing the ion transfer capillary are two seperate things, for the cleaning of the spray chamber, simply lower the source temperature in the method file to 0 ( will take about 10 minutes to cool down to a reasonable range), then put the instrument in standby and open and clean the spray chamber.
For the ion transfer capillary, you have to vent the system.
Agilent has a new cleaning solution apart from using the wire and cotton technique.
Simply follow the instructions for cleaning the ion transfer capillary, its a simple wash procedure, you use the alconox detergent i believe and perform a set of sonications in this detergent as well as di water.
The advice I can give is to be gentle when taking the capillary in or out, you dont want to damage the 1500$ delicate item.
Also, even more important, remember that things are easy to take off, make sure you remember how to put things back the way they were
After the cleaning, pump down, and after an overnight stabilization of the system, your tune abundances should look awesome!
Best of luck.