HILIC frustration
Posted: Fri Mar 22, 2013 9:57 am
Experienced a lot of ion-exchange on the ZIC-HILIC column so I changed to the Waters XBridge HILIC column. But there seems to be a lot of ion-exchange on this one as well.
I run my samples with a water gradient (5 to 60% water), but some peaks never elute (mostly excipients such as methionne). That would be OK, but the column gets overloaded with excipients after a few injections, causing retention time drift and pressure increase.
After risning the column with 0.5M NaCl, the column is as new again. But to be able to use HILIC I need to run wash programs between every injection. The other option would be to use more buffer in the mobile phase, but the only option seems to be ammonium acetate (completely destroying the sensitivity at 220 nm).
Just some frustration written down... If you know of any HILIC column without ion-exchange or how to shut the ion-exchange down (in 95% acetonitrile) - please let me know
I run my samples with a water gradient (5 to 60% water), but some peaks never elute (mostly excipients such as methionne). That would be OK, but the column gets overloaded with excipients after a few injections, causing retention time drift and pressure increase.
After risning the column with 0.5M NaCl, the column is as new again. But to be able to use HILIC I need to run wash programs between every injection. The other option would be to use more buffer in the mobile phase, but the only option seems to be ammonium acetate (completely destroying the sensitivity at 220 nm).