RP-HPLC baseline drift?

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

9 posts Page 1 of 1
Hi
Recently I have observed relatively high baseline drifting on our Agilent 1100 system. I'm running a linear gradient of 5-90% Acetonitrile in water (0.1% TFA). Initially the tubing was 0.10mm but I changed it to 0.17mm, which somewhat helped on issue. As I only observe the drifting at 215nm and not 280, I suspected it could be related to the ACN, however exchanging it with a fresh bottle didn't solve it, neither does it seem to be the column, as I have tried with 3 different columns. The inserted pictures is of a run where the sample is a bit too concentrated, but the baseline at 215nm is drifting up to approx. 350 mAu.
Image

280nm
Image

Any help appreciated
Erik
Baseline drift with TFA / ACN systems is almost unavoidable because of the interactions between TFA and ACN and the effects on the absorbance spectrum of TFA:
http://www.chromatographyonline.com/lcg ... rticle.pdf
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Some people use 0.1% TFA in eluent A and 0.05% TFA in eluent B. This approach tends to reduce - if not eliminate - the upwards drift but it might infuence the separation due to the pH rise during the gradient. But it might not be an issue here judging from the separation task simlicity in this particular case.

Best Regards
Learn Innovate and Share

Dancho Dikov
Thanks Tom , quite comprehensive paper! I had a closer look at the degasser module, and it might here the issue arises, since one of the vacuum tubes had tenderised. I ordered a replacement and hopefully that'll help.

Danko I'll try your advice and only use 0.05% TFA in the ACN. But what do you mean by "pH rise" during gradient? Will the pH not decrease with an increasing eluent B percentage if you only have half the concentration of TFA in eluent B?

Thanks again
Erik
Hi Erik,
Will the pH not decrease with an increasing eluent B percentage if you only have half the concentration of TFA in eluent B?
Less acid = higher pH

Not very significant in this case though. So you don't need to worry. just try and see :-)

best Regards
Learn Innovate and Share

Dancho Dikov
yes yes yes of course, it's me not thinking.
Hi again
So now I have exchanged the damaged vacuum tubing, but the baseline drifting still persist - no difference at all. I'll try the 0.05% TFA in eluent B and see if that helps, although the issue must have occurred due to some change in the system, it wouldn't just occur out of the blue!

\Erik
If it doesn't help I'll buy you a beer - or two :wink:

Best Regards
Learn Innovate and Share

Dancho Dikov
I did help as stated, thanks again.

\Erik
9 posts Page 1 of 1

Who is online

In total there are 16 users online :: 3 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot], Google [Bot], Google Adsense [Bot] and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry