Varian 1200 GC-MSMS

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

4 posts Page 1 of 1
I hope someone can give me an advice. My Ion source gets dirty very fast. I have to clean the ion volume almost every week and the ionblock every 3 weeks or so. We test pesticides and make a cleanup before we measure. The ionblock didnt look so yellow after 3 month before the problem began. The only thing i can think of is a leak so oxygen can come in. Im going to change the column and the fittings.

DId someone have the same problem ?
... The only thing i can think of is a leak so oxygen can come in. Im going to change the column and the fittings.

DId someone have the same problem ?
Before changing column and fittings execute test for leaks. You will get some information where the leak exist (provided it is a leak).
I already did that and there is no leak detektable. That's why i waited so long to finally do that. maybe the column is to old and starts leaking.
I know what happend now, the tiny plastic cap on the transferline was broken.
4 posts Page 1 of 1

Who is online

In total there are 22 users online :: 0 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry