I remember I just ran carrier gas through the brand new column but had not conditioned the column after installation. After that, I used the column to ran a few batches of other gradient analysis with max temp up to 320 deg C (with short hold time). Plus another batch of ethylene glycol quantitation with max temp up to 300 deg C (hold time 5 min).
Is the poor peak shape due to failure to condition the column after installation?
I changed new liner, septum, cut 12 cm column head and cleaned ion source to make sure other variables are optimal. Then condition the column again at 280 deg C 3 hours, then 300 deg C 1.5 hr. Injection between each conditioning showed tiny improvement.
Can I improve the column condition now by conditioning for longer period of time?
Please kindly give your expert comments.
Thank you.