Iam trying to optimize 1,2,4-Triazole (T) on Agilent 6460 MS/MS detector.
Solvents/flow are A 90% H2O 0.01% Formic acid, 5mM Amonium formate
B 10% ACN 0.01% Formic acid
Column: Short C18 (20mm) 0.3 ml/min
When i inject 10ug/ml 1,2,4-Triazole in mobile phase i get nice peak
When i inject 10ug/ml 1,2,4-Triazole in ACN/H2O 1:1 i get negative peak no signal for 70m/z
Any idea why ?
Thanks for help
I would think that your problem is too much organic in the injection solution. If you have too high of a concentration of organic you will not be able to retain compounds on the column. If you need to inject this matrix, then you need to find a way to dilute the sample down before it hits the column. When I encounter this problem on a Waters system, my trick is to inject into a much larger sample loop than what would be required for the injection volume. This way the sample gets diluted with the mobile phase before it hits the column.
2 posts Page 1 of 1

Who is online

In total there are 21 users online :: 0 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 21 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry