Injection of Saline Solution into LC-MS

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

5 posts Page 1 of 1
Does anyone have any thoughts as to the feasibility of injecting 0.9% saline solutions into an LC-MS? The injection volume will be somewhere in the range of 10-25uL, analytical column is a C18 with 1% acetic acid methanol/water mobile phase. I realize salts are generally frowned upon for LC-MS, but the concentration of the chemical I am needing to detect in the saline is already very low and I'd rather not dilute if this isn't a problem. Thanks for any advice!
10-25ul should not be too much of a problem when injected into the mobile phase, it should dilute it enough. Just be sure not to make too many injections without checking your source for salt buildup. I remember one of the old Agilent adds showing a source after multiple injections of undiluted sea water samples with the salt crusted on the source and the system still working, though I would not want to let mine get in such a shape.
The past is there to guide us into the future, not to dwell in.
It may not be too harmful but if possible I would still avoid it. I can think of 2 options:

-use a diverting valve that will direct the mobile phase to the waste until the solvent peak has eluted.

-use a C18 SPE step prior to LC-MS which would allow you to get rid of the salt and possibly concentrate your sample by evaporation of the eluting solvent.
I routinely analyze environmental samples of salt water and do everything to minimize the salt concentration, from diluting all samples when you have the sensitivity and always using a diverter valve. However, there are times when the sample can't be diluted further and I have analyzed straight salt water (usually with 25-50% organic added). I would not usually employ SPE solely to get rid of the salt. I would matrix match your standards if you are shooting samples with high salt content and keep a close eye on your source, especially API 3000 types.
jmmunson wrote:
Does anyone have any thoughts as to the feasibility of injecting 0.9% saline solutions into an LC-MS? The injection volume will be somewhere in the range of 10-25uL, analytical column is a C18 with 1% acetic acid methanol/water mobile phase. I realize salts are generally frowned upon for LC-MS, but the concentration of the chemical I am needing to detect in the saline is already very low and I'd rather not dilute if this isn't a problem. Thanks for any advice!


Ditto on the use of valve switching. As I said in another thread, pick a switch time later than the void but prior to the beginning of your first peak's elution. Salts will usually come out with the void.

If you have access to a syringe pump you can use post-column addition to visualize whether there is matrix suppression in the region where your peaks elute.

I also have to say - 1% acetic in your mobile phase is quite strong and I think it's likely to cause ion suppression. Normally I'd not go higher than 0.1% formic or acetic, although in one case with HILIC-MS I found 0.5% formic to be necessary for peak shape.
5 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry