Hi! I don't have a lot of experience with GC maintenance. I just replaced the column and ran a few blanks through it. Blanks look good, but once I ran a standard through it, the baseline doesn't seem to settle.

It's a 60 m column, id: 0.250 mm, film: 0.15 um. DB-23

With a column this long, I'm going to condition it over night. (Whereas prior to this, I conditioned for 3.5 hours at a high temperature.) Hopefully the baseline settles at this point.

Would love any other suggestions? Thank you! :D
I like to condition my columns without connecting the detector end. Install it in the injector, make sure your flow rate is good, and make sure there are no leaks. Bring the injector up to temp. Heat the oven up to the max temp the column allows (I believe this one is 240) for a few hours (the longer, the better) and then cool it down to your initial temp. Now check to see if the injector ferrule is leaking and tighten it if necessary. Connect the detector end and when leak free, heat the detector to your desired temp. Heat the oven up to max temp again for 10 minutes or so. Cool it back down to your initial temp and make sure the injector and detector ferrules aren't leaking. New ferrules will expand when heated and contract when cooled so this is a break in for them. What you're seeing now may be due to a leak, not bleed. Good luck and please let us know how it goes.
Please remember there are two maximum temperatures for any column.

One is for isothermal analysis. The other is for programed temperatures where the final temperature is not held constant.

Do not subject the column for an extended time at the programmed max temperature, as the expected lifetime of the column will be shortened.

best wishes,

Rod
What are your blanks, and what is in the standard ? When you say that the baseline does not settle, does it rise steadily during a run or is it unstable before the injection ?

When did you last change the inlet liner and septum ?

I would advise against a lot of "conditioning" until other possible causes of baseline drift (if that is indeed what you are seeing) have been eliminated. If you have a leak or dirty carrier gas then holding the column at high temperatures for long periods can only damage it.

Peter
Peter Apps
I like to condition my columns without connecting the detector end. Install it in the injector, make sure your flow rate is good, and make sure there are no leaks. Bring the injector up to temp. Heat the oven up to the max temp the column allows (I believe this one is 240) for a few hours (the longer, the better) and then cool it down to your initial temp. Now check to see if the injector ferrule is leaking and tighten it if necessary. Connect the detector end and when leak free, heat the detector to your desired temp. Heat the oven up to max temp again for 10 minutes or so. Cool it back down to your initial temp and make sure the injector and detector ferrules aren't leaking. New ferrules will expand when heated and contract when cooled so this is a break in for them. What you're seeing now may be due to a leak, not bleed. Good luck and please let us know how it goes.
For anyone new to doing column maintenance there is a trick here if you are using vespel/graphite ferrules that can save some headaches with this problem. Keep extra ferrules in a small beaker inside the oven(not in their plastic case of course) and the temperature cycling will pre-shrink the ferrules so that when they are used you will get little if any loosening after installation.
The past is there to guide us into the future, not to dwell in.
Thanks James,

That sounds like a good plan.

Peter
Peter Apps
6 posts Page 1 of 1

Who is online

In total there are 17 users online :: 0 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry