Empower 2 data acquisition issue
Posted: Tue Feb 26, 2013 11:53 pm
by joshkn
I'm having trouble acquiring data from a 2420 ELSD. I can control the instrument and monitor the baseline just fine in Empower, but when I hit inject no data trace appears, and I get "data missing" in the review. This system works well with the PDA. There are no related messages in the message center.
Any ideas out there?
Josh
Re: Empower 2 data acquisition issue
Posted: Wed Feb 27, 2013 1:54 am
by DR
Start cable hooked to detector?
Posted: Wed Feb 27, 2013 7:32 am
by Bull76
I'd agree with DR on this. You need an inject start from the ELSD to the Alliance. You'll also need a stop flow from the ELSD to the Alliance also if it is not already in place. This will help prevent the drift tube from getting flooded.
Posted: Wed Feb 27, 2013 9:14 pm
by joshkn
Thanks! - the additional connections fixed the issue.
EDIT: Well it works for the first injection - then on subsequent injections there is no signal :(
Posted: Wed Aug 07, 2013 9:45 pm
by ncairns
I'm getting exactly the same problem, first injection works then ELSD remains on time 0.00 and will not start for the second or any subsequent injection - did you find a solution??
Posted: Thu Aug 08, 2013 8:38 am
by aceto_81
So your start worked, now a stop cable?