I'm having trouble acquiring data from a 2420 ELSD. I can control the instrument and monitor the baseline just fine in Empower, but when I hit inject no data trace appears, and I get "data missing" in the review. This system works well with the PDA. There are no related messages in the message center.

Any ideas out there?

Thanks,
Josh
Start cable hooked to detector?
Thanks,
DR
Image
Hi,

I'd agree with DR on this. You need an inject start from the ELSD to the Alliance. You'll also need a stop flow from the ELSD to the Alliance also if it is not already in place. This will help prevent the drift tube from getting flooded.
Thanks! - the additional connections fixed the issue.

EDIT: Well it works for the first injection - then on subsequent injections there is no signal :(
I'm getting exactly the same problem, first injection works then ELSD remains on time 0.00 and will not start for the second or any subsequent injection - did you find a solution??
So your start worked, now a stop cable?
6 posts Page 1 of 1

Who is online

In total there are 14 users online :: 2 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot], Google [Bot] and 12 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry