Indeed, the cation exchange columns can last a long time. The problem is that the handling requirements are different enough from C18 that many people don't have their intuition developed for ion exchange. To make the column last you should:
* Observe the maximum pressure and flow specifications. Even a brief exposure to excess flow can cause a head space.
* Avoid organic solvents. They can cause a slow pressure buildup.
* Inject clean, filtered samples dissolved in water.
* Avoid surfactants in your sample if possible.
* Use a guard column.
* Take measures agaings iron contamination. It can kill the chromatography, but it is easy to remove.
* Passivate your HPLC system
* Avoid stainless steel filters in your solvent reservoirs. They corrode too easily.
This advice is aimed at the fully sulfonated PS-DVB gel that is typical for glyphosate analysis (BioRad, Pickering, etc.).
One last point. The ion exchange capacity of these columns is large, and the mobile phase has low ionic strength. Equilibration is very slow.
Hi Mark,
what is passivate?
Can I use peek material for the mobile phase filter?
Should the sample loop and all other tubings be of Peek material?