Tips on getting ES-MS detector response for small molecules?

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

5 posts Page 1 of 1
I have an LC-MS in ESI mode. The problem I'm having is that I'm not getting a response from the MS detector for most analytes. The mass range for my analytes (the known ones anyway) is 144-154. I'm seeing peaks on the DAD but only noise on the MS. Any thoughts? Do I have to switch to CI?

I have an agilent 1100 HPLC with G1946D single quad in ESI and I'm running reverse phase separation with water/acetonitrile and a C18 column.
Incitor Inc.
"Making renewable fuels a reality"
http://www.incitor.com
I'm no mass spectrometrist, but (in very general terms) ESI ionization essentially occurs in solution (inside the solvent droplets), so it's probably a good idea to have *some* buffer in the mobile phase. Also, not all molecules are sufficiently acidic/basic to ionize under ESI conditions, in which case APCI may be required.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
I'm no mass spectrometrist, but (in very general terms) ESI ionization essentially occurs in solution (inside the solvent droplets), so it's probably a good idea to have *some* buffer in the mobile phase. Also, not all molecules are sufficiently acidic/basic to ionize under ESI conditions, in which case APCI may be required.
Thanks for the tip. Yeah, I've thought about that, I will give it a try and see if that works. If not, I'm going to have to switch to APCI.
Incitor Inc.
"Making renewable fuels a reality"
http://www.incitor.com
what are the analytes? Not everything ionises well in ESI. Also remember that if they're very small and volatile, they will dry off just like the solvent does...
what are the analytes? Not everything ionises well in ESI. Also remember that if they're very small and volatile, they will dry off just like the solvent does...
My analytes are not that volatile. They remain after a rotovap.

Actually, I added 0.1% v/v acetic acid and it really improved the signal from the MS. I'm going to try formic acid next to see if it does any better.
Incitor Inc.
"Making renewable fuels a reality"
http://www.incitor.com
5 posts Page 1 of 1

Who is online

In total there are 41 users online :: 2 registered, 0 hidden and 39 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Google [Bot], Majestic-12 [Bot] and 39 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry