Figure 1: Before bakeout. Isooctane with BHT. Temperature program: 170 deg C for 0 min, then 0.6 degC/min to 207.5 deg C for 0 min, then 5 deg C/min to 225 deg C for 0 min. Column: ZB wax 100% PEG.
______________________________________________________________________________________________________________________

Signal: FID1 A, Front Signal
RT AREA NAME
37.241 294.5889 EPA/DHA Std-#1
46.667 528.2255 EPA/DHA Std-#1
56.836 3321.5303 EPA/DHA Std-#1
Sum 4144.3447
Figure 2: Before bakeout. EPA, DHA, and C23:0 ISTD. Temperature program: 170 deg C for 0 min, then 0.6 degC/min to 207.5 deg C for 0 min, then 5 deg C/min to 225 deg C for 0 min. Column: ZB wax 100% PEG.
______________________________________________________________________________________________________________________
Figure 3: After bake out (Oven temp 240, inlet @ 325, & det @ 325). Chromatogram of Isooctane with BHT. Temperature program: 170 deg C for 0 min, then 0.6 degC/min to 207.5 deg C for 0 min, then 5 deg C/min to 225 deg C for 0 min. Column: ZB wax 100% PEG.
____________________________________________________________________________________________________________________

Signal: FID1 A, Front Signal
RT AREA NAME
36.586 290.2884 EPA/DHA Std-#1
46.032 73.2491 EPA/DHA Std-#1
56.018 187.3622 EPA/DHA Std-#1
Sum 550.8997
Figure 4: After bake out (Oven temp 240, inlet @ 325, & det @ 325). Chromatogram of EPA, DHA, and C23:0 ISTD. Temperature program: 170 deg C for 0 min, then 0.6 degC/min to 207.5 deg C for 0 min, then 5 deg C/min to 225 deg C for 0 min. Column: ZB wax 100% PEG. Peak areas have greatly improved from those in Figure 2. The peak area values are similiar to the peak areas I was getting before the shutdown occured.
______________________________________________________________________________________________________________________
Figure 5: Blank run-No injection. Temperature program: 170 deg C for 0 min, then 0.6 degC/min to 207.5 deg C for 0 min, then 5 deg C/min to 225 deg C for 0 min. Column: ZB wax 100% PEG.
_____________________________________________________________________________________________________________________
I disconnected the column from the detector and set detector at 450 deg C and oven at 230 deg C. Baseline was stable at 0 pA. Anymore advice on what is going on here

Thanks for your help !!!!