The real answer to this question should be directed to the QA/QC Officer of your company; however, for practical purpose, the answer is yes, especially for changes #1 and #2.
You will see a significant change in retention time, resolution, increased sensitivity, etc. when moving from a packed column to a capillary column, and if the phase is polar, possibly a change in elution order of the compounds.
Difference in retention time, resolution, and increased sensitivity is self-explanatory; however, change in elution order is not. It has to due to difference in the polarity of the phase at different temperatures and the different amount of phase coating from a packed column compared to a capillary column. There is no direct method of getting the same phase load on a capillary column converting from a packed column; therefore, the actual polarity may not be the same.
If using a polar phase, the phase polarity changes with changes in temperature. This is also true if there is a difference in flow rate, or for capillary columns, linear velocity. For example, if using a method published in a journal article or a manufacturer of capillary columns, and the flow rate or linear velocity you use is different from the article method, compounds will elute at a different temperature. As stated above, polarity of polar phases change with temperature, which can create elution order changes. This is one of the reasons that it is very important to use individual compound standards to set up and achieve optimum chromatographic conditions.
As for possibly doing a mini validation, I would suggest doing a completely new validation. Mini validations are typically used when changing from one brand of column to another, but still the same phase type and dimensions.
I hope this helps answer your questions, but again, you should check with your QA/QC Officer and the SOP for method validations.