by
lmh » Fri Jan 18, 2013 3:54 pm
Perhaps some explanation useful for the less-experienced? (sorry if this is patronising!):
(1) Outgassing. Gas can only come out, if it's already in. If you've used a good online degasser, then there shouldn't be much dissolved gas in the aqueous and acetonitrile phases, and when they're mixed, although the solubility limit may be less than it was, the amount of gas that's present is still soluble. This is particularly true at the very high pressures the pump is creating.
(1b) But yes, purge your lines. Gas can redissolve back into the solvent lines, through the walls of the tubing, between the degasser and the pump. Purging also gets rid of any bubbles that have managed to appear in the last week that the instrument wasn't used...
(2a) Sudden pressure changes: a sudden pressure change is bad mechanically for a column, if the particles can move in response to it, creating voids. Modern well-packed columns with spherical particles are pretty much optimally in place (a bit like marbles arrange themselves on a tray), so a sudden pressure change won't mess them up. Older columns with irregular silica could never pack that way, so they were more vulnerable. Of course some people are still using irregular silica today, and these people, together with those using other less usual packing materials, should be more careful than the rest of us.
(2b) Any sudden pressure change is bad, but sudden decreases don't really happen so easily, because the pump doesn't encourage back-flow. If the pump stops, the pressure in the system has to be relieved by liquid flowing away through the column, and generally the system will have a pulse-dampener and some elasticity to be relieved. On suddenly starting the pump, it would be possible to raise the pressure more quickly if the pump is quick enough, but again it depends on the capability of the pulse-dampener.
(3) Pumping with pre-mixed solvents for the start and end of the gradient. A gradient pump has a certain accuracy and precision. Maybe when you specify 25% it could be anywhere from 24-26. If you want to run a (rather silly) gradient from 24% to 26% and you do this by putting 0% on channel A and 100% on channel B, and specifying 24-26 in your gradient table, you could actually be running a gradient anywhere between 23 and 27, and you don't really know whether it spans 2%. If you put premixed 24% on A and 26% on B, and specify in the gradient table 0-100%, the pump's +/- 1% precision is now suddenly improved 50-fold, because a 2% error as it passes through x% is now 2% of the difference between 24% and 26%...
(3b) But this depends on how well you mix your samples, and relies on solvent bottles not standing around long enough to evaporate significantly (which changes the mix). Labs who are good at mixing, and who habitually run the same gradients, will probably pre-mix. Labs who run a huge variety of wildly different methods are likely to use 100% ACN and rely on their pump in all but the most critical situations. No one wants to be throwing away half-used batches of expensive ACN-based mixes, or using buffers made up weeks ago.