Had a question with regards to results on my Chromatography results
I'm using Shimadzu LC 20AT LPG pump with SIL-20HT Autosampler with SPD-M20A PhotoDiode Array detector with CTO-20 Column Thermostat and CBM-20 Controller
I have a Sunscreen method that runs Isocratic based on following conditions
Column : Kinetex Phenomenex column 2.6 micro PFP 100 x 4.6mm
Column Temperature: 45 degrees C
Mobil Phase: 60 ACN: 40 Water
Flow Rate : 1 ml/min at 313 nm
Back Pressure: Steady at 98 Bars
Injection: 10 microliter
Im running 3 sunscreens
Octylmethoxycinnamate or Octionoxate
Avobenzone
Octisalate
Here is the problem
I'm developing this method to get all the peaks in ten minute run
for the entire last week, I ran couple of sequences and I get near to perfect calibration- I'm running 5 point calibration
But my results do change- ( I'm making the same sample amount )
Monday - I had Octinoxate and Avobenzone at right amount but the Octislate was low
Octinoxate should be about 6% and Aveobenzone is about 2% and Octisalate must be around 4.5%
I got close results to two of the analytes - by trying to dissolve the samples in 50/50 Acetonitrile -Water- It worked for two of the analytes but not the third one( Octisalate)
The standard prep is in 9:1 Ratio Methanol: Water. I can't make my standards in 60:40 ACN: Water simply because the standards are not dissolving correctly- So, I though there must be something wrong with my sample prep.
I took a different approach
I dissolved the samples first in 1 part Water and then QS them with 4 part MeOH to the mark.
Now Im getting the Avobenzone and Octisalate in right amounts and Octionxate in 7% plus region.
Please note, that I have perfect calibration and I beleive my sample amounts are correct as well
Here is how I made the first set of samples and Octisalate was out of range
I dissolved them in 50:50 Water/ Acetonitrile and vortexed the samples in 25 ml vol flask. after seating down for some time I did a 1/10 dilution ( 50:50 ACN:Water) and injected them
The second way how I prepared the samples was
I dissolved them in Water first (5 ml) and then vortexed the samples to disperse them thoroughly- After that I QS the samples with 20 ml Methanol in 25 ml flask and vortexed them again very good , so there will not be any undissolved samples left.
And proceeded with 1/10 dilution
The second set of dilutions were with 9:1 ratio Methanol: Water for the 10 ml flasks and I injected them
And this time Octisalate and Avobenzone were in range but Octionxate was out of the range reading around 7.0%
At this point I'm suspicious of my sample preparation. I know I should make my samples just like standards , but sometimes this doesn't help with the method.
Retention times are coming fine with the Methanol: Water mixes
Standard Retention Times with (60:40 ACn: Water) are as follows
Octinoxate at 4.44 minutes
Octisalate at 5.39 minutes
Avobenzone at 6.5 minutes
Samples with matter of seconds are back and forth- but no significant shift in retention times.
From what I know and read, I know that I have to make my samples weaker than mobile phase- I'm not following that , I know- But I have a problem with dissolving all the standards. SO, thats why Im dissolving my standards in 9:1 MeOH: Water solution and my samples in 8:2 Methanol: Water solution.
If you guys have any ideas please share ;
Thanks in Advance