connecting a coulochem III detector

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

4 posts Page 1 of 1
I have an Agilent 1200 HPLC that we want to connect a Coulochem III detector to. At the moment the LC and an e-SAT/IN box are connected to a Waters LAC/E box. The Coulochem III is connected to the e-SAT/IN. Both the LC and the e-SAT/IN are controlled through Empower 2.

I don't have the Coulochem connected to the 1200 LC because I don't know how to do so. I was told by someone at ESA (Thermo) to connect the out start signal from the autosampler to the coulochem. We have a G1367C HiP-ALS autosampler. There are no "out start signal" ports on the back. We have RS-232, Remote, and LAN/CAN ports as well as two CAN DC-out pin plugs that I can't figure out what they are supposed to do.

At the moment we can get the SAT/IN to monitor the baseline using the Empower 2 and we get a signal from the detector. I haven't tried a manual injection while it was monitoring because I'm not sure how to do that without stopping the flow. No, I don't want to set up a secondary injector valve to test it.

When I start an injection, the SAT/IN goes into a ready state but doesn't receive a signal to start collecting data. (The data light never starts blinking which indicates that it's collecting data) During the sample runs, I see the potentials in the cells change on the coulochem display but no data is collected in Empower.

I realize there is an Aglient general purpose cable that goes from the Remote on the autosampler to whatever you plug it in to. Is this the right cable to use? Or is there one from ESA or Waters that I have to use? Do I connect the cable to the SAT/IN box or the detector? And if I do get the cables right, what are the settings supposed to be in Empower 2? There are no "remote signal relay settings" in the instrument setup interface.

The above might sound like a stupid question but dammit Jim I'm a chemist not an electrician!
Did you ever find out how to do this? I am trying sometning similiar but need to know how to send the start signal from our Agilent autosampler to the ESATIN module.
I have an Agilent 1200 HPLC that we want to connect a Coulochem III detector to. At the moment the LC and an e-SAT/IN box are connected to a Waters LAC/E box. The Coulochem III is connected to the e-SAT/IN. Both the LC and the e-SAT/IN are controlled through Empower 2.

I don't have the Coulochem connected to the 1200 LC because I don't know how to do so. I was told by someone at ESA (Thermo) to connect the out start signal from the autosampler to the coulochem. We have a G1367C HiP-ALS autosampler. There are no "out start signal" ports on the back. We have RS-232, Remote, and LAN/CAN ports as well as two CAN DC-out pin plugs that I can't figure out what they are supposed to do.

At the moment we can get the SAT/IN to monitor the baseline using the Empower 2 and we get a signal from the detector. I haven't tried a manual injection while it was monitoring because I'm not sure how to do that without stopping the flow. No, I don't want to set up a secondary injector valve to test it.

When I start an injection, the SAT/IN goes into a ready state but doesn't receive a signal to start collecting data. (The data light never starts blinking which indicates that it's collecting data) During the sample runs, I see the potentials in the cells change on the coulochem display but no data is collected in Empower.

I realize there is an Aglient general purpose cable that goes from the Remote on the autosampler to whatever you plug it in to. Is this the right cable to use? Or is there one from ESA or Waters that I have to use? Do I connect the cable to the SAT/IN box or the detector? And if I do get the cables right, what are the settings supposed to be in Empower 2? There are no "remote signal relay settings" in the instrument setup interface.

The above might sound like a stupid question but dammit Jim I'm a chemist not an electrician!
Does the ALS have a contact closure accessory board installed? If so, that can be used to time program an electrical closure or to supply voltage at the time you choose.
I am not familiar with a contact closure board and am not sure if one is installed or not. There is a remote output and according to the manual one of the pins is designated as a signal start. The ESATIN has inputs for the signal start that are both a positive and a negative. I am not sure how to go from one pin to a positive and a negative.

Thanks
Does the ALS have a contact closure accessory board installed? If so, that can be used to time program an electrical closure or to supply voltage at the time you choose.
4 posts Page 1 of 1

Who is online

In total there are 16 users online :: 0 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry