Particle surface coverage of LC column

Posted: Wed Dec 05, 2012 1:36 pm
by JI2002
Could anyone give some insight on how to determine the surface coverage of the particles, either by experiment or by calculation? Thanks.

Re: Particle surface coverage of LC column

Posted: Wed Dec 05, 2012 4:51 pm
by tom jupille
You need to know the surface area (usually determined by porosimetry) and the percentage carbon (standard CHN analysis).

Convert the carbon percentage to moles (use the appropriate molecular weight for your ligand) and divide by the surface area.

Re: Particle surface coverage of LC column

Posted: Thu Dec 06, 2012 1:03 am
by JI2002
Thanks. That helps a lot. Is the CHN analysis the same method the column manufactures use to determine the carbon load?

Re: Particle surface coverage of LC column

Posted: Thu Dec 06, 2012 1:28 am
by tom jupille
Pretty much (actually, only the "C" part). Basically, burning off the organic part and measuring the resulting CO2.