Hi guys, I have a 5973MSD and it does not show any signal, not even a baseline. I tuned it and it works fine when I tune it. Everything else is fine with the system. I do not get any error messages. Any suggestions? Thanks in advance
The baseline cut off is over 9000!
-
- Posts: 693
- Joined: Mon Mar 28, 2005 5:38 pm
I had that but the MS wasn't getting a start signal so it wasn't collecting data. The service guy had left it unplugged.
If the MS is telling you that it is acquiring data but you don't see anything, check that your tune file in the method is the same one you were adjusting, ie that you are using the same filament.
If the MS is telling you that it is acquiring data but you don't see anything, check that your tune file in the method is the same one you were adjusting, ie that you are using the same filament.
-
- Posts: 17
- Joined: Fri Oct 19, 2012 8:00 pm
Well, I just checked it and there is nothing wrong with the connections. The MS runs the tune correctly which should mean that there isn't anything wrong with the MS itself. When i run the sampling program, it does not switch on which is indicated by the lack of baseline signal. Also, i checked the sampling program and i have set all the parameters correctly. It was working fine a week ago. So its strange that its malfunctioning now when nothing has been changedI had that but the MS wasn't getting a start signal so it wasn't collecting data. The service guy had left it unplugged.
If the MS is telling you that it is acquiring data but you don't see anything, check that your tune file in the method is the same one you were adjusting, ie that you are using the same filament.
The baseline cut off is over 9000!
-
- Posts: 1
- Joined: Thu Dec 20, 2012 7:59 am
In the MS Scan acquisition settings, check the value for Threshold (an acquisition rejection filter). Try a low value, even 0. If it is set too high, it can reject entire spectra, leading to no baseline.
4 posts
Page 1 of 1
Who is online
In total there are 41 users online :: 1 registered, 0 hidden and 40 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Majestic-12 [Bot] and 40 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Majestic-12 [Bot] and 40 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science