Dear everyone,

I have a problem with my HPLC RI system, (both with SE, eluent water with 0,3% NaCl; and NP, eluent water).
The baseline seems to be steady before the run, and during every single run the baseline jumps 5-10 times from 0 to -50..-80 µRIu

Image

https://docs.google.com/file/d/0B4f2VbJ ... VjeU0/edit

Thanks for any help!
Dear everyone,

I have a problem with my HPLC RI system, (both with SE, eluent water with 0,3% NaCl; and NP, eluent water).
The baseline seems to be steady before the run, and during every single run the baseline jumps 5-10 times from 0 to -50..-80 µRIu. The pump pressure and oven temperature are both stable and okay.

Here is how it looks:

https://docs.google.com/file/d/0B4f2VbJ ... VjeU0/edit

Thanks for any help!
Air bubble in the flow cell?
Problems with the reference cell?

PeterS
Thank you very much, I have tried one hour detektor purge with methanol now, and its not better, so I guess flow or reference cell contamination or air bubbles can be excluded...

Any other ideas?
That link didn't work (page not found).

However, spikes that occur only during a run are usually symptomatic of air bubbles (or, far less common, particles making their way out of the column).
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Hello,

Maybe flushing with methanol is not enough. Years ago we had similar problems and we had to wash the cell with 1M NaOH, water and propanol.
Then our signal became stable.

PeterS
Dear everyone,

I have a problem with ...
And the question is why you ask about HPLC in GC forum ?
Thank you very much for the reply, but it does not look like air bubbles.

This looks like: the baseline is on zero for 10 min, than on - 50 uRIU for 5 min, than back to zero, than -80 uRIU for 7 minutes again, back again, and so on, and so on.

Thanks for any help!
Okay, thanks for the advice, I'll try it. But it really doesn't look like air bubbles, because when the basline drops dramatically to - 50 uRIU it also stays there for 5-10 minutes, and only than returnes to zero.



ps: To answer Mr. Proper as well: Of course it was by accident, that I have used the GC forum at first, but apparently it was a good choicce: I have got more answers by now, than on the HPLC forum.



Thanks
Threads now merged in appropriate forum.
Thanks,
DR
Image
...
apparently it was a good choicce: I have got more answers by now, than on the HPLC forum.
Anna, it wasn't. You simply violated forum rules. If you were more patient and waited one or two working days you would get answers in HPLC forum without the need of "cleaning" by Mod.
OK, dear, I was joking by saying: "it was a good choice", because in the same sentence I have already told that it was by accident.

I really feel offended by being charged with violation of the rules, and having to apologize for so many times about an obivously unintended and ridiculously small-scale mistake.

So I say thank you to those who were helping, and I say farewell to this forum.

Anna
12 posts Page 1 of 1

Who is online

In total there are 13 users online :: 1 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 12 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry