waters has introduced some chemistries with 2.5um particles, has any one used this, any feed back.
or any information about other brands with this particle size.
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
You're right, I'm also quite happy with the Kinetex. Astonishingly high efficiencies at a reasonably low backpressure. Nevertheless I'm not so happy with Phenomenex' marketing claim "UHPLC perfomance on any HPLC system". According to my experience, you won't get those high efficiencies by just putting a Kinetex in a conventional HPLC system. You MUST optimize the system (smaller volume flow-cell, optimized capillary connections) in order to exploit these columns. Even then, it's better to stick with 3.0 or even 4.6mm ID columns, as the extra-column band-spreading with 2mm columns will be to high on these systems.I'm becoming increasingly convinced by solid core particles. I've been trying out Kinetix 2.6u, which is solid core and therefore comparable to smaller particles while keeping the reasonable back-pressure. Phenomenex aren't lying: it really is as good as the 1.7u particles from the point of view of resolution, and yet I can run my 50mm*2mm easily at 400 uL/min on a conventional LC system.
Full ACK. I usually scale the injection volume down according to the changed column diameter for a first try. Depending on the result I'd further optimize injection volume, solvent and sample concentration.I wasn’t thinking of solvent effects.
The thing is, if you load a certain sample amount on 3 mm column - for instance – and if it’s the optimum load for that column, then switching to a column with narrower diameter – 2 mm for instance – will necessitate sample amount reduction in order to avoid sample overload. Sample overload can often be confused (looking at the chromatogram) with extra column effects.
Injection volume reduction may, or may not be necessary. Depends on several factors and is needed mostly when sample solvent is a stronger solvent than the mobile phase, while the opposite (weaker solvent) is often beneficial.
When talking about 250x4,6mm monsters, ACKIn my experience, column overloading in an analytical scenario is rarely a problem compared to solvent effects. I'd be more concerned about going beyond the linear range of the detector than overloading the column in most situations.