lmh » Thu Nov 15, 2012 11:36 am
I think what's happening is that Chemstation reverts to the flow conditions/solvent mix in the "isocratic" section of the pump control at any time before the gradient starts, and after the gradient is finished.
I'm not sure there's a way to stop it doing this, but what you could try is putting the flow to zero in the flow box at the top right, and specifying the actual flow you want, only in the gradient table. I haven't tried this, but I think that would mean that chemstation would pump 100% acetonitrile (or whatever you specify) at zero flow until it's ready to start the gradient. It would then run the gradient, and when it's finished, revert to zero flow, so although technically the solvent mix is wrong, none of it is reaching the column. The down-side would be a pressure disturbance before your first real run. In my case I always run a blank that I throw away as my first sample. It wastes instrument time, but it means all subsequent injections start with the column in the same defined state.