Hi
I using ICS 3000 but I am getting poor precision. I have flushed the syringe to get rid off air bubbles but the RSD is > 15%. Can you help? Thanks
If you're using an AS3500 autosampler be sure to have the injection type specified to FULL in the program file.
Make sure the flush solution is degassed high purity water.