Hi,

I had to chang the rotor seal in the injection Valve and now I need to startup the whole HPLC system to get the air out of it. Does anyone know how to proceed or can give me a link? I have been looking for a long time but found no description how to proceed.

The HPLC system is from SHIMADZU with the following components:

- LC-20AD pump
- RID-10A detector
- DGU-20A3 solvent degasser
- CTO-20A oven
- 6 position injection valve (Rheodyne) and an Aminex column

I would be very glad if anyone could provide help!
Simply , disconnect the column inlet , change the rheodyne valve to inject position , start the pump and adjust the flow rate to 3 ml/min , wait 10 seconds and connect the column again .
Sorry , I forgot to mention :
Adjust the flow rate to appropriate value before connecting the column.

3ml/min flow is too high for analytical column.
Thank you very much!
4 posts Page 1 of 1

Who is online

In total there are 57 users online :: 0 registered, 0 hidden and 57 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 57 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry