Waters Alliance 2690 needle bottom sensor malfunction

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
We are getting a needle bottom sensor malfunction message with the injector of a Waters Alliance 2690 HPLC system. There was no evidence of the needle being bent from hitting the bottom of the vials but for good measure the needle was recently changed and its travel set 2mm above maximum depth. However, the message still occurs every few hours of operation. Rebooting the Alliance system restores normal function for a few hours. Has anybody else experienced this problem and if so is it rectified by replacing the injector sensors? TIA, Tony.

I think you have to clean up & lube the screws that drive the needle assembly to correct that error message. Evidently the motor(s) that drive the needle are sensitive to any changes in the torque required to move the carriage and error out with little provocation.
Thanks,
DR
Image
2 posts Page 1 of 1

Who is online

In total there are 12 users online :: 1 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 11 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry