(1) Inject a solvent blank, which is used to dissolve the standard in (2).
(2) Inject a standard with known concentration.
(3) Inject a sample (i.e. dissolve the unknown substance in the same solvent as in (2).
(4) Compare the chromatograms obtained in (1), (2) and (3).
(5) Result intepretation: Peaks appear in chromatograms (1) and (3) will be disregarded as they are believed to be generated by the solvent or mobile phase. Refer the analyte peak to chromatogram (2). Peaks only appear in sample chromatogram other than analyte peak and will be treated as impurity peaks.
(6) Purity calculation: Determine peak area percentage of peak analyte in sample chromatogram.
Can any one of you experts give me some advice on the above method design?