- Posts: 8
- Joined: Mon Nov 07, 2011 1:50 pm
I am working with a GC/MS Agilent 6890-5973 and I have problems with the optimisation of the repeller and the entrance lens in the manual tuning (problem occured after source maintenance). I can't get the Gaussian shape that I should have normally (see picture below).

This problem also occurs in the atune.u mode although the Tune report is fine (parameters are ok, abundances are ok...). I would have started my analysis if it wasn't that I was in front of the screen when the tune was running and I saw quickly that the repeller did not give a Gaussian shape..
I tried cleaning and reinstalling the source and got similar results... By the way, all the wires are connected to the right places and there are no loose connections. I changed the repeller, the insulators, the draw-out plate and even the electron multiplier (because it was overdue) and not of it worked...
I switched the source of the 5973 with the one in my other GC/MS (5975) and got good results on the second GC/MS (and bad results with the good source on the malfunctioning GC/MS). So, the problem is not the source itself.
I though that it might be a corrupted file so, after rebooting my entire system, I loaded a previous tune, saved it under a different name and ran it and still got bad results. Now, the only option I see is that this is an electronic problem. Maybe it could be the ceramic board on which the wires are connected, I've heard that sometimes it fails but I don't know what it does exactly when it's broken...
Does anybody have any idea what is the problem or what could I do next?
Thanks!