On Waters 2489,2487 2996 and 2998 detectors how should I test for detector noise and also how often should this be preformed.

Is this a required by the FDA?

Thanks
This should be a standard part of your performance qualification (PQ). Yes, it is required, but how often is a matter of your company's SOPs. I would say annually as a minimum and any time the detector has been worked on.

If you use Empower, you can check it:
http://www.waters.com/webassets/cms/sup ... ower_2.pdf

There is also an ASTM practice: E1657-98
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Thanks for your help Tom