I am testing pestcides in Rice according to COMMISSION REGULATION (EU) No 893/2010.In this there is a combined MRL for Carbedezim and benomyl since benomyl has low stability and is hydrolysed to carbendezim easily.I have doubt that is benomyl & Carbendezim included in the mix standard or i run carbendezim only.
-
- Posts: 310
- Joined: Thu Jun 18, 2009 7:59 pm
You run carbendazim only. There are special prep methods that you can use to keep from hydrolyzing the benomyl, but under the standard EU/FDA prep procedures you will only get the peak for MBC.
Mark Krause
Laboratory Director
Krause Analytical
Austin, TX USA
Laboratory Director
Krause Analytical
Austin, TX USA
-
- Posts: 61
- Joined: Fri Sep 02, 2011 1:26 am
Thanks for your valuble infomation
3 posts
Page 1 of 1
Who is online
In total there are 44 users online :: 0 registered, 0 hidden and 44 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 44 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 44 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science