josebenjamin,
The thing to keep in mind with columns of this sort is that they are a bit on the fragile side so cleaning protocols should be "gentle" but this does not they can't be effective. But, patience is needed since these columns have quite high loading capacity and it can take overnight or even a couple of days of cleaning to get all the garbage off a heavily fouled column. Thus, it's a good idea to target the cleaning with a bit of knowledge as to the nature of the contamination.
If the column pressure is high, it could be due to particulate based plugging of the inlet. In this case your best hope is to replace the inlet frit. If this doesn't reduce the column pressure, it's likely that the top of the column is contaminated with particulate. Just about your only hope in this case is to reverse the flow but make sure though column doesn't have any headspace before you do this, otherwise the column is ruined (although if the column is more or less dysfunctional at this point, there is little risk). Another possible cause of high pressure is contamination with "swelling agents". Such swelling agents include organic solvents and other organic molecules which might have a high solubility in the stationary phase and limited solubility in water. The cleaning protocol in this case is dependent upon the nature of the contaminant. If the contaminant is non-ionic or possibly a weakly acidic organic, the best strategy is to use an organic solvent rinse. 30% acetonitrile will generally not cause irreversible damage to such columns but it may take a day or two to clean the column and restore it back to original aqueous eluents. Keep in mind that such columns generally don't care for abrupt solvent changes. A gradient to and from solvent containing mobile phases is advisable. If the contaminant is possibly a cationic species, the mobile phase will also need to contain a substantial amount of acid in order to facilitate elution of the compound (although such compounds generally cause bed shrinkage and headspace and often such damage is unrecoverable). Transition metal contamination is best removed with a chelating agent which is acid soluble such as oxalic acid. Can you comment a bit more on possible sources of contamination?