Have a small air leak? Check your calibrant vial.

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

4 posts Page 1 of 1
I thought I'd share.

I have been running at 2.5% N2 and ~5% H2O on my 6890-5975 for as long as I can remember. Yesterday I refilled the calibration vial. I am now running at 0.2% N2.
Thanks for the tip. I'm sure this is a spot that is commonly forgotten about as calibration solution is rarely added.
~Ty~
This happened to me recently. I blame my technical services guy, as he refused to put but a few drops in the bulb when the instrument was installed.
Time flies like an arrow. Fruit flies like a banana.
You can tell if you have air and H2O in the cal gas vessel by turning the cal gas off when looking at air and H2O in profile mode. If the air and H2O area counts stay the same when cal gas is on and off, you have a leak in the system. If the air and H2O area counts drop when the cal gas is turned off, it is probably in the cal gas vessel.
4 posts Page 1 of 1

Who is online

In total there are 110 users online :: 0 registered, 0 hidden and 110 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 110 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry