How is the best condition for HPLC room (%RH, temperature)?
What is the effect of room’s circumstance to HPLC? Any literature is pleased.
Best regards,
SYX
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
HPLC is used for both normal and reverse phase. the organic solvents used are volatile and change in the conc. of mobile phase may take place at higher temperature. This factors should also be kept in mind.
Have you ever worked with HPLC instrument kept at temperature condition 35C (e.g. uncontrolled room temperatures in summers in countries like Africa).
I have experienced the change in retention time due to Change in mobile phase composition with time.
Thus conditions of temperature and %RH of HPLC room should be maintained at certain levels.
What about moist absorption by mobile phase components in reservoir, especially for unaqueous system. May it need special cap with dehumidifier (moist absorber) that connected to vent hole of the resevoir? Does anyone know the supplier of this kind of cap?No, the volatility of the solvent is irrelevant for a change in composition, since you are heating the solvent under high-pressure conditions. Volatility of solvents may only be important for creating gas bubbles in the detector, but this can be fixed as well, if necessary.